For highest accuracy, the value sent with the CAL command must exactly
equal the actual output value of the adjustment source. It is recommended
that 10V be used for CAL 10 and 10 K ohms be used for CAL 10E3.
NOTE: Any standard value between 1 V and 12V or
1 K and 12 K can be used. A value less than 10V or less than 10 K
will introduce additional uncertainty to the multimeter's accuracy
specifications. For example, a 1 V DC standard can be used instead of
10V (you would execute CAL 1.0000). A 1 K standard can used instead
of 10 K (you would execute CAL 1E3). Each case degrades the
accuracy specifications of the instrument.