the Meter Input range +/- 12.5V,input impedance > 10^9 OHM。But the non - linearity error after fitting will still great . The Temperature coefficient is very low but only because of my real-time aoto "Self-calibration" design both the hardware and software。
You saied your meter's Performance good that < 1 ppm nonlinearity and <1 ppm/C temperature coefficient, I hoped to see more Test data and the Test equipment should be more reliable。Or,zhe ADS1282 it's very high-performance than other ADCs?Even so, the input attenuation circuit(16V to 5V or 2.5V)' error is inevitable,How do you killed those problems?
Test equipment - Datron 4000A calibrator and two Solartron 7081 DMMs, modded to LTZ1000 voltage references.Input Amplifier is bootstrapped and have a metal-foil resistors divider and MAX44246 buffer. Schematics is similar to Datron 1281 DMM.